Gary Bamberger
Hi Devan,
The program that you attached won’t even compile. Example:
S4,
IF N>=5 [@True, @False]
@True: Z^Z_Sucrose ---> S5
@False: ---> S2
This is not valid code and if I tried to compile it, it would say illegal or missing input statement. There are several locations in the program where the same problem exists.
This makes it a little hard to give advice because you could not possibly be using this program to run your experiments.
Now having said that it looks like you are trying to use a PHM-100VSS pump. The pump is only turned on when a Z^Z_Sucrose pulse is issued. In S.S.1, S4 & S7 this Z-pulse is issued. In S10, S13, S16, S19, S22, S25, S28, and S31 no Z-pulse is issued.
Hopefully this is the cause of the problem. If that doesn’t help, then the next step is to see if you can turn on the pump when using MED Test. If it doesn’t work in MED Test it will never work when running your MED-PC program.
Please let me know if this helps solve your problem.
Gary